Topic: Acuson X300 Power SupplyPosted: Sep 15 2011 at 6:34am |
|
Hi.
I have a Acuson X300 power supply problem. When i try to power up the system two situations occur: 1. The leds OCP and AC light up and the system doesn´t boot up 2. The led FANS lights up and the system doesn´t boot up Most of the time point 2 occurs when i try to boot up the system. Best regards Paulo |
|
![]() |
|
caruso122
Admin Group
Joined: Apr 09 2009 Location: Noblesville, IN Points: 85 |
Post Options
Thanks(0)
Quote Reply
Posted: Sep 15 2011 at 1:42pm |
|
Hello Paulo,
Have you tried measuring the power supply outputs? If they are low or missing, try pulling out each board in the card cage to see if one is causing the power supply to crowbar.
If that doesn't help, I would try replacing the supply.
Regards,
Chris
|
|
|
C&C Medical Solutions, Inc.
_____________________ Christopher Turner Phone: 317-219-3616 x103 Fax: 317-770-0560 Yahoo IM: caruso122 Skype IM: cturner122 Email: chris@ccmedicalsolutions.com |
|
![]() |
|
fanie
Newbie
Joined: Oct 13 2009 Points: 28 |
Post Options
Thanks(0)
Quote Reply
Posted: Nov 17 2011 at 5:44am |
|
Hello Chris
Regarding the X300 power supply, i found the DC/DC power supply with problems, causing the leds on the USPA-AC power supply to light up. Having repaired the DC/DC power supply, on power up the system hangs on the message "Checking system configurations.." All the voltage tests are O.K. What could be the problem? Best regards Paulo Ferreira |
|
![]() |
|
caruso122
Admin Group
Joined: Apr 09 2009 Location: Noblesville, IN Points: 85 |
Post Options
Thanks(0)
Quote Reply
Posted: Nov 17 2011 at 9:23am |
|
Hi Paulo,
It sounds like either the operating software has been corrupted or a board is now bad due to the power supply issue.
I would re-load software first. After that if you still have an issue, reseat all the boards in the card cage one by one to see if the error message goes away. If that does not correct the issue, try booting the system with a minimum configuration as outlined in the service manual.
Best regards,
Chris
|
|
|
C&C Medical Solutions, Inc.
_____________________ Christopher Turner Phone: 317-219-3616 x103 Fax: 317-770-0560 Yahoo IM: caruso122 Skype IM: cturner122 Email: chris@ccmedicalsolutions.com |
|
![]() |
|
fanie
Newbie
Joined: Oct 13 2009 Points: 28 |
Post Options
Thanks(0)
Quote Reply
Posted: Nov 17 2011 at 11:10am |
|
Hi Chris
Thanks for your rapid response to my issue. I have re-loaded the software, but still stuck on message "Checking system configurations". I don´t have the service manual, could you please indicate `to me which boards are needed for minimum configuration? Regards Paulo Ferreira |
|
![]() |
|
caruso122
Admin Group
Joined: Apr 09 2009 Location: Noblesville, IN Points: 85 |
Post Options
Thanks(0)
Quote Reply
Posted: Nov 17 2011 at 11:18am |
|
Hi Paulo,
Unfortunately, I do not have the service manual for the X300, otherwise I would gladly supply you with that information.
Best regards,
Chris
|
|
|
C&C Medical Solutions, Inc.